I have never had this issue with the deadweight tester
Are you using all the right type of instruments?
When I do the test I follow this procedure......
First: You need the following recently calibrated instruments - deadweight tester, voltage source, and DMM. Remember that the accuracy of your calibration is only as good as the total uncertainty of this "system." (Note: Some deadweight testers can be calibrated for your local gravitational constant! There are also other types of pressure generating equipment available, but they all depend on a known-good gauge/transducer.)
Second: I use an 11 point test: 0% 20% 40% 60% 80% 100% 80% 60% 40% 20% 0%. Slowly rotate the piston at each step so stiction doesn't affect your reading. Going up and down on the pressure will check for hysteresis, especially important with gauges.
Third: Keep good records and check on a regular basis.
Be very careful when comparing the accuracy of pressure transducers! There appears to be no standard convention so different manufacturers will state different numbers which are difficult to compare.
Let us know if you can't get a reading, I do have another way